CapeSym CSI-14x25c-SIPM-T and GS-USB-PRO
Posted: 23 Jul 2022, 13:38
Hi All,
The subject of Silicone photomultipliers has come up several times recently and I wasn't sure how to deal with it, so I invested in a tiny CapeSym CsI detector just to have a play with.
The model number is CSI-14X25C-SIPM-T
https://capescint.com/product/macropixe ... 5c-sipm-t/
(No association with the seller.)
Unlike regular PMT's the SiPM only needs around 30 Volts bias, so connecting it directly to the GS would kill it for sure, but I came up with a simple hack 😉
Due to the 1M internal load resistor in the GS you can bridge the output with a 100K resistor, this results in a voltage drop of around 90% so 300V becomes ~30V.
Since the SiPM also draws a bit of current I had to turn up the voltage on the GS to just over 400V but that worked almost perfectly. I say "almost" because the small 100K pull down resistor I used slowly warms up, causing upwards voltage drift. Thinking of ways to prevent or compensate for this, maybe a 30V zener diode in parallel with the resistor could work.
Only three of the pins on the CapeSym are used, the other 4 are exclusively for the connection of a digital temperature probe. I suspect the actual detector will also suffer some drift as the temperature changes.
Other than that, the only change I made was to replace the recommended 200 ohm load resistor from anode to ground with 2K ohm, this has the effect of making the pulse longer.
As you can see, the signal connects to the BNC input just like any 2 wire detector.
Important to note, unlike normal PMT's the pulse from the SiPM is positive and the GS inverts the pulse, so in PRA you need to set the Threshold to a negative LLD, (suggest -0.5)
The resolution of this little detector is great, were it not for the drift it would be in the range 5.3% @ 662, but as it drifted a bit over a 10 minute run it slowly increased to 6.3%.
The next stage would be to make up a little probe handle containing these parts with a socket for the CapeSym. probably with an SHV and a BNC so connection to the GS becomes the same as any other probe.
Ultimately I don't see too much of a barrier to making my own SiPM detectors, other than the raw cost of the actual SiPM chips.
Steven
The subject of Silicone photomultipliers has come up several times recently and I wasn't sure how to deal with it, so I invested in a tiny CapeSym CsI detector just to have a play with.
The model number is CSI-14X25C-SIPM-T
https://capescint.com/product/macropixe ... 5c-sipm-t/
(No association with the seller.)
Unlike regular PMT's the SiPM only needs around 30 Volts bias, so connecting it directly to the GS would kill it for sure, but I came up with a simple hack 😉
Due to the 1M internal load resistor in the GS you can bridge the output with a 100K resistor, this results in a voltage drop of around 90% so 300V becomes ~30V.
Since the SiPM also draws a bit of current I had to turn up the voltage on the GS to just over 400V but that worked almost perfectly. I say "almost" because the small 100K pull down resistor I used slowly warms up, causing upwards voltage drift. Thinking of ways to prevent or compensate for this, maybe a 30V zener diode in parallel with the resistor could work.
Only three of the pins on the CapeSym are used, the other 4 are exclusively for the connection of a digital temperature probe. I suspect the actual detector will also suffer some drift as the temperature changes.
Other than that, the only change I made was to replace the recommended 200 ohm load resistor from anode to ground with 2K ohm, this has the effect of making the pulse longer.
As you can see, the signal connects to the BNC input just like any 2 wire detector.
Important to note, unlike normal PMT's the pulse from the SiPM is positive and the GS inverts the pulse, so in PRA you need to set the Threshold to a negative LLD, (suggest -0.5)
The resolution of this little detector is great, were it not for the drift it would be in the range 5.3% @ 662, but as it drifted a bit over a 10 minute run it slowly increased to 6.3%.
The next stage would be to make up a little probe handle containing these parts with a socket for the CapeSym. probably with an SHV and a BNC so connection to the GS becomes the same as any other probe.
Ultimately I don't see too much of a barrier to making my own SiPM detectors, other than the raw cost of the actual SiPM chips.
Steven