Struggling with LYSO
Posted: 29 Oct 2017, 02:43
I recently made a probe for gamma spectroscopy with a LYSO crystal matrix (5x5 crystals, each 4x4x20mm, for a total of 20x20x20mm cube) and a Hamamatsu R9420 tube (FWIW both the LYSO and the tube were bought from Tom at Irad on ebay). The problem I have is that the LYSO has an intrinsic background due to the presence of 176Lu which does emit a number of low-energy gammas in its decay cascade.
However, I did not expect the background to be that high. I'm getting approx. 140000cpm with no gamma source nearby! (and some lead shielding). Apparently it's normal for this type of crystal. And since the source of the gammas is inside the crystal itself, the counts are very high even though the crystal when counted with an external detector (a 3inch BC412-based one) only gives around 5-6000cpm.
I have a 137Cs source (10uCi) which I'm trying to use to calibrate this new detector, but have some trouble with background subtraction. Here are below some of my most recent (and best so far) attempts.
The parameters I'm using are: 1000V tube bias, 100mV rejection (there are a LOT of pulses below 100mV, X-rays mostly which I'm trying to eliminate from the equation), 2mV/channel, range 2V (=1000 channels). This is not with a Theremino-like adapter but with a special software-controlled voltage source and acquisition is with a Picoscope2000 USB oscilloscope after a signal conditioning amplifier. Both are driven by a custom software interface.
I get (among others) an output as text files containing a listing of channel numbers and number of pulses counted in the channel during the experiment. I did some text processing with bash and awk to obtain test minus background (with/without the 137Cs source) then render graphs in LibreOffice. I noticed that the counts appeared to be a little shifted between channels between experiments so I also did a manual shifting of one of the file counts (the background) by 1 row to better fit the data. I didn't correct negative values but rather left them in place as-is. Shifting the background file values by 2 places down did not improve things, quite the contrary.
It's still not very neat. All counts were over 10 minutes.
I am concerned mainly with the maximum peaks of the lutetium and cesium mostly overlapping... And they shouldn't.
However, I did not expect the background to be that high. I'm getting approx. 140000cpm with no gamma source nearby! (and some lead shielding). Apparently it's normal for this type of crystal. And since the source of the gammas is inside the crystal itself, the counts are very high even though the crystal when counted with an external detector (a 3inch BC412-based one) only gives around 5-6000cpm.
I have a 137Cs source (10uCi) which I'm trying to use to calibrate this new detector, but have some trouble with background subtraction. Here are below some of my most recent (and best so far) attempts.
The parameters I'm using are: 1000V tube bias, 100mV rejection (there are a LOT of pulses below 100mV, X-rays mostly which I'm trying to eliminate from the equation), 2mV/channel, range 2V (=1000 channels). This is not with a Theremino-like adapter but with a special software-controlled voltage source and acquisition is with a Picoscope2000 USB oscilloscope after a signal conditioning amplifier. Both are driven by a custom software interface.
I get (among others) an output as text files containing a listing of channel numbers and number of pulses counted in the channel during the experiment. I did some text processing with bash and awk to obtain test minus background (with/without the 137Cs source) then render graphs in LibreOffice. I noticed that the counts appeared to be a little shifted between channels between experiments so I also did a manual shifting of one of the file counts (the background) by 1 row to better fit the data. I didn't correct negative values but rather left them in place as-is. Shifting the background file values by 2 places down did not improve things, quite the contrary.
It's still not very neat. All counts were over 10 minutes.
I am concerned mainly with the maximum peaks of the lutetium and cesium mostly overlapping... And they shouldn't.