CSI Detector

Scintillation crystals, PMTs, voltage dividers etc...
Zagros
Posts: 18
Joined: 23 Jul 2017, 01:53

CSI Detector

Post by Zagros » 08 Aug 2019, 05:27

Hello to all!
Anyone knows this detector?


https://rover.ebay.com/rover/0/0/0?mpre ... 4313963951

Thanks

User avatar
iRad
Posts: 217
Joined: 01 May 2015, 12:27

Re: CSI Detector

Post by iRad » 08 Aug 2019, 07:11

Based on the model number "13B38/0.3-CS-X", it appears to be a (X) customer custom 13mm diameter by 38mm long CsI (probably Tl) crystal canned together with a TO-8 Metal Package PMT (probably a R5600 or R7400). The pins are very tiny and fragile.

I'm sure Luuk could tell you more about it.

luuk
Posts: 313
Joined: 01 May 2015, 21:00

Re: CSI Detector

Post by luuk » 09 Aug 2019, 01:29

Hi Nicola,
The detector is exactly as Tom described it, 13mm round, 38mm long, the crystal is CsI(Tl), the pmt R7400u.
It is difficult to connect them because the pmt wires are very fragile and it is hard to find the right pmt socket.
If it works you expect a resolution from around 7.2-7.5%@662KeV average.
Often these secondhand detectors have broken pins(wires) or the pmt is broken.
Luuk

Zagros
Posts: 18
Joined: 23 Jul 2017, 01:53

Re: CSI Detector

Post by Zagros » 09 Aug 2019, 05:02

luuk wrote: ↑
09 Aug 2019, 01:29
Hi Nicola,
The detector is exactly as Tom described it, 13mm round, 38mm long, the crystal is CsI(Tl), the pmt R7400u.
It is difficult to connect them because the pmt wires are very fragile and it is hard to find the right pmt socket.
If it works you expect a resolution from around 7.2-7.5%@662KeV average.
Often these secondhand detectors have broken pins(wires) or the pmt is broken.
Luuk
Ok.
Thank you!
I’ll look for something else.
Bye!

User avatar
iRad
Posts: 217
Joined: 01 May 2015, 12:27

Re: CSI Detector

Post by iRad » 12 Aug 2019, 11:43

These Scionix CsI detectors are nice for their small size, however they use one of the more difficult sockets to find and/or wire a voltage divider for. Wiring directly to the very fragile pins is possible, but risk to the PMT is great and is not advised.
You do not have the required permissions to view the files attached to this post.

User avatar
iRad
Posts: 217
Joined: 01 May 2015, 12:27

Re: CSI Detector

Post by iRad » 12 Aug 2019, 14:14

Here is what wiring the small TO-8 PMTs socket is like... I wired this one for most efficient power use. Notice in the pictures it now runs well at only 650 volts.
You do not have the required permissions to view the files attached to this post.